Default AVO Valve tester question.

Does someone know:

I have an AVO MK3 tester.
The highest current range for the meter is 100mA.
A few years ago I fixed a 200W Marshall guitar amp for a friend, but I hit a problem when it came to checking the output valves(4XKT88). The test data states an anode current of 130mA which is obviously off the scale. Is there a way of increasing the range of the current meter or is it just not possible for bigger valves to be tested on the MK3 ?

I meant to ask on here at the time but forgot all about it.

Regards, Phil.
__________________
Commission for Dark Skies (CfDS)

Last edited by Phil Y; 17th November 2012 at 04:44 PM.
Reply With Quote
  #2  
Old 17th November 2012, 09:02 PM
Phil Y's Avatar
Phil Y Phil Y is offline
Senior Member
 
Join Date: Jul 2006
Location: Bristol, UK
Posts: 816
Default Re: AVO Valve tester question.

Please ignore the above, I have gone back through the paperwork I got with the tester (loose photocopies)and found a second set of settings with lower anode and screen voltages that give a usable anode current.
One set must be for a different model.

Phil.
__________________
Commission for Dark Skies (CfDS)
Reply With Quote
  #3  
Old 18th November 2012, 01:00 AM
Richard Richard is offline
Super Moderator
 
Join Date: Dec 2005
Location: Notts
Posts: 5,357
Default Re: AVO Valve tester question.

Phil,

I have characteristics of

-20V grid, 250V anode, 250V screen, 57mA IA and 8mA/V for KT88

I also use them for 6550 as the "book" specifies characteristics including 400V anode which my machine doesn't have but it's a very similar valve and tests fine with KT88 settings.

So it's worth thinking of a similar valve if you get stuck.

The other thing you can do is test a known good valve at settings that are available on your machine and as near to the recommended ones as the machine allows. The you can then use those settings for other valves to see how they compare with the good one. I do this for 300B as the "book" ones are too high neg grid volts for my machine so have found my own which suit the machine.
Reply With Quote
  #4  
Old 18th November 2012, 07:01 PM
Phil Y's Avatar
Phil Y Phil Y is offline
Senior Member
 
Join Date: Jul 2006
Location: Bristol, UK
Posts: 816
Default Re: AVO Valve tester question.

Hello Richard,
Those settings for KT88 are the same as the new ones I found but thanks for confirming them.
Both your suggestions for getting around testing problems are good ideas, I will remember them for future use.

Thanks, Phil.
__________________
Commission for Dark Skies (CfDS)
Reply With Quote
Reply


Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off
Forum Jump


All times are GMT +1. The time now is 05:08 AM.


Powered by vBulletin® Version 3.7.3
Copyright ©2000 - 2024, Jelsoft Enterprises Ltd.
Copyright World Designs